Hello Forum,
We just converted a CR23x program to a CR3000 and noticed that the variable names don't use a consistent case.
For example, the FSL from the CR23X was "AIR_150_AVG" and the output from the CR3000 is "AIR_150_Avg". The programmer is not specifying _AVG in the Alias definitions.
Our QC scripts are case sensitive so they see these two as different variables, but in fact, they are the same.
Is there a setting somewhere that I can tweak to make the table name suffixes from the CR3000 default upper case?
Hello,
You can use the FieldNames instruction to change the name. Here's a simple program. The second Average in the data table shows the use of fieldnames. The first Average will keep the default name, but the second will assume a new name:
Public Tref, TCTemp1, TCTemp2
DataTable (Table,True,1000)
DataInterval (0,60,Sec,10)
Average (1,TCTemp1,FP2,False)
Average (1,TCTemp2,FP2,False)
FieldNames("TCTemp2_AVG")'<<<<<<<<Rename output
EndTable
BeginProg
Scan (1,Sec,3,0)
PanelTemp (Tref,250)
TCDiff (TCTemp1,1,mV2_5C,1,TypeT,Tref,True ,0,250,1.0,0)
TCDiff (TCTemp2,1,mV2_5C,2,TypeT,Tref,True ,0,250,1.0,0)
CallTable Table
NextScan
EndProg
Dana
Thanks Dana,
Do you know if there is a way to define new field names if reps is more than 1?
ie:
Public Tref, TCTemp(2)
Alias TCTemp(1) = TEMP_150_CM
Alias TCTemp(2) = TEMP_250_CM
DataTable (Table,True,1000)
DataInterval (0,60,Sec,10)
Average (2,TCTemp,FP2,False)
FieldNames("TEMP_150_CM_AVG", "TEMP_250_CM_AVG" )'<<Would this work?
Hello,
Fieldnames for multiple outputs needs to be one quoted string for all reps; i.e.,
FieldNames("Name1, Name2")
as opposed to
FieldNames("Name1", "Name2")
Dana
Hi,
Is it possible to pass a public string into the FieldNames() function? For example:
///
Public PTemp, batt_volt, TcoupleTemp, ExampleData(40) As Float,
Public i,
Public fieldNameString = "ExampleString,ExampleString2"
DataTable (ExampleDataTable,1,-1)
Sample (1,batt_volt,FP2)
Sample (30,ExampleData(),IEEE4)
FieldNames(fieldNameString)
EndTable
///
The result of this code is fieldNameString as the first Example Data column header, not ExampleString.
All I'm looking for is a way to evaluate the argument used by the FieldNames() function before it's used by the FieldNames() function.
Hi Johnathan, Dana,
I'm trying to do the same thing as in the last post, passing a string into the FieldNames() function.
Const siteNumber = "123"
FieldNames("Battery_"+siteNumber)
Is this possible?
Cheers